Project
How to Check and Calibrate a Humidity Sensor
November 29, 2017 by Raymond Genovese
How accurate is your humidity sensor? Find out with this project.
Humidity sensors are commonplace, relatively inexpensive, and come in many different varieties. Too often, we check the datasheet, use them with an interface, and (as long as the values “look reasonable”) we accept the results.
In this project, we demonstrate how to go a step further and verify the accuracy of a humidity sensor. We also illustrate a general method for sensor calibration and apply the method to calibrate the results to improve the accuracy of the humidity measurements.
Testing setup used in the project (left to right, Quark D2000 microcontroller board, sensor interface, HIH5030 sensor in a micro-environment).
Project Fundamentals
To check the accuracy of a sensor, obtained values are compared to a reference standard. To check the accuracy of a humidity sensor, we use the “saturated salt” method to produce the standards. Put simply, certain salts (i.e., ionic compounds such as table salt or potassium chloride), when dissolved in an aqueous solution, produce an atmosphere of a known humidity (see reference PDF).
These chemical properties are used to create micro-environments of known relative humidity (RH) percentages (i.e., reference standards), and the sensors are read inside the micro-environment. Specifically, we will make a solution in a sealed jar to preserve the atmosphere and then place the connected sensor in the sealed jar. Subsequently, the sensor is repeatedly read and the values recorded.
By repeating the procedure using several different salts, each producing a different relative humidity, we can develop a profile for the sensor under test. Since we know what the relative humidity is for each micro-environment, we can assess the deviations of our sensor readings from those known values, and thus, evaluate the accuracy of the sensor.
If the deviations are substantial, but not insurmountable, we can apply mathematical calibration procedures in software to increase the accuracy of the measurements.
A Word about Safety
Before going further, it is essential that you handle the chemicals used in this project responsibly.
  • Read the safety data sheet (SDS, or sometimes MSDS (material safety data sheet)) for each of the chemicals used (links for the SDS for each salt used are provided below, and you can also conduct literature searches on each salt and their safe handling procedures).
  • Do not inhale or ingest the chemicals.
  • Do not let the chemicals contact your skin or eyes (use gloves and goggles).
  • Do not prepare the solutions in the same area that food is prepared.
  • Properly store the chemicals.
  • Properly discard the solutions and all of the instruments used to prepare the solutions so that exposures do not take place accidentally.
  • Before starting, know what to do if an accidental exposure takes place (see the safety datasheet).
Salts Used
In general, the more RH atmospheres that you can produce for reference standards, the better the characterization of the sensor under test will be. There is, however, always a limit on resources in a practical sense. In this project, four reference standards were used and the salts used to produce the reference standards were chosen to cover a range of possible RH values, but also with consideration to safety, availability, and cost.
The salts below were chosen. In the case of sodium chloride (table salt), pure kosher salt was obtained cheaply at a local grocery store. If you go that route, avoid using table salt with additives, such as iodine or anti-caking agents.
Salts Used in the Project
Salt % RH (at 25°C) Source Safety Data Sheet
Lithium Chloride 11.30 Home Science Tools SDS for LiCl
Magnesium Chloride 32.78 Home Science Tools SDS for MgCl
Sodium Chloride 75.29 Various (see text) SDS for NaCl
Potassium Chloride 84.34 Home Science Tools SDS for KCl
Creating a Micro-Environment
We have standards for nearly everything and there is even one for creating a stable RH from an aqueous solution (see ASTM E104 - 02(2012)).  While my bench, and probably yours, is not an official testing laboratory, it is worthwhile to follow the specifications in the standard as closely as you can.
Note also that the results presented in this project, while collected with care, should not be construed as reflecting or indicating an overall quality statement of the accuracy of any brand of sensor. Only a small number of sensors were tested and those used had different ages and different usage histories.
For each salt, a slushy mixture was created by adding distilled water to a consistency similar to very wet sand. Four or five tablespoons of chemical and one tablespoon of distilled water can be tried, but you may have to do a little experimenting.
The mixture was made in a small jar with a tight seal. Glass or even plastic should work well, so long as it can keep the atmosphere inside. A small hole can be made in the top of the jar to run connecting wires to the sensor interface and then to a microcontroller. The connected sensor is then positioned approximately 0.5-1.0 inch above the mixture. Take care that the sensor never directly contacts the solution or it will likely be damaged. To hold the connection in place and to seal the hole in the cap, some easily removable contact putty can be used.
It is important that you allow plenty of time for equilibration before you take the final reading. I tested this issue empirically, taking readings every minute for up to six hours in selected test cases. In my experience, this was longer than needed and I settled on 90-120 minutes equilibration time for each sensor and salt. Then an average of the last five readings were used for the final value. For all cases, the five values showed very little, if any, difference.
Additionally, all readings were taken at about 25° C (± 1°) ambient temperature, and the RH value used for each standard was that listed for 25° C (see this PDF for the values).
Sensor and Micro-environment
HIH5030 sensor on a carrier board inside a micro-environment containing sodium chloride.
Hardware
Microcontroller
In this project, we interface the sensors using a Quark D2000 microcontroller. The D2000 is a 3V board with I2C and analog-to-digital interfaces. For more information on the D2000, the reader is directed to these previous AAC articles:
Keep in mind, though, that most any other microcontroller with the appropriate interfaces can be used.
Sensor Interfaces
Sensors tested in the project; A) HIH8121, B) HIH5030, C) DHT-22 (AM2302), D) HIH6030 (on a carrier board).
Four different types of humidity sensors were tested: DHT-22 (two were used), HIH5030, HIH6030, and HIH8121.The schematics below illustrate the simple interfaces used for each type of sensor, and consultation with the linked data sheets will provide background information for the circuits.
  • The DHT-22 is a temperature and humidity sensor with a proprietary serial output (see this AAC article for more information on the serial protocol used).
  • The HIH5030 is a humidity sensor with analog (voltage) output. The interface for this sensor uses an op-amp in a unity-gain configuration for impedance matching.
  • The HIH6030 and HIH8121 are temperature and humidity sensors that use the I2C protocol (see this AAC article for more information on the I2C communication procedures used).
DHT-22 to D2000 interface.
DHT-22 BOM: U1, DHT-22 sensor; R1, 4.7kΩ resistor; C1, 0.1 µF capacitor.
HIH5030 to D2000 interface.
HIH5030 BOM: U1, HIH3050 sensor; U2, MCP601P op-amp; C1, 1.0 µF capacitor; C2, 0.1 µF capacitor.
HIH6030 to D2000 interface.
HIH6030 BOM: U1, HIH6030 sensor; R1 and R2, 2.2 kΩ resistor; C1, 0.22 µF capacitor; C2, 0.1 µF capacitor.
HIH8121 to D2000 interface.
HIH8121 BOM: U1, HIH8121 sensor; R1 and R2, 2.2 kΩ resistor; C1, 0.22 µF capacitor.
Sensor Software
All of the programs for gathering sensor data are written in the C language and can be downloaded by clicking on the “Humidity Sensor Project Code” button. Each is commented and straightforward. For each sensor, the program simply reads the sensor every minute and sends the value to a serial monitor. As such, they should be easy to adapt to your particular application.
Screenshots of output from DHT22.c (left) and HIH5030.c (right).
Program files for the project can be downloaded by clicking the link below.
Project_Code.zip
Sensor Evaluation Procedure
The table below contains the data from evaluating the sensors in each of the four micro-environments.
Percent RH for the Test Sensors (OBS = observed value, ERR = error as difference from standard, RMSE = root mean squared error)
  DHT #1 DHT #2 HIH5030 HIH6030 HIH8121
Reference RH OBS ERR OBS ERR OBS ERR OBS ERR OBS ERR
11.30 (LiCl) 12.56 1.26 16.29 4.99 13.02 1.72 20.79 9.49 12.31 1.01
32.78 (MgCl) 32.36 -0.42 33.79 1.01 33.46 0.68 40.77 7.99 32.43 -0.35
75.29 (NaCl) 73.04 -2.25 74.50 -0.79 77.74 2.45 83.83 8.54 76.63 1.34
84.34 (KCl) 82.30 -2.04 82.15 -2.19 85.84 1.50 93.43 9.09 85.01 0.67
RMSE   1.657   2.799   1.708   8.796   0.920
Once you have collected the data from the sensor performance in stable environments of known relative humidity, you can numerically evaluate a sensor’s accuracy.
Note that in the table, we calculated the error for each sensor at each RH standard. We can’t, however, simply average those values to evaluate the sensor because some values are positive and other values are negative. If we simply took an average, the resulting value would minimize the average error since the positive and negative values would cancel each other.
Instead, we calculate a root mean square error (RMSE) to characterize the sensor’s accuracy. The formula for RMSE is below:
where O is the observed sensor value and I is the ideal sensor value (i.e., the reference standard). To calculate RMSE, we square each error (the deviation from the reference standard), then calculate the arithmetic average of those values, and finally, take the square root of the average.
Once you have characterized the accuracy of the sensor, you can use the RMSE to decide whether it is necessary to calibrate the sensor. In some cases, the RMSE is small and completely acceptable for your application and you can reasonably decide that no calibration is required.
For example, the results for the HIH8121 are impressive. The RMSE is less than 1% and all sample points have an error less than 2%.
On the other hand, in some cases, you may find that the sensor response is so poor and irregular that you simply decide that another sensor is required for your application.
The decision to calibrate should always take into consideration the degree of accuracy necessary for the task. Nevertheless, we can improve the accuracy of the sensor readings by calibration, for all of the sensors in the table.
Sensor Calibration Procedure
To calibrate a sensor, we need to first mathematically determine the function that relates the ideal values to the observed values. A linear regression procedure can be used to determine that function.
The word “linear” in the name of the regression procedure does not mean a linear function. Instead, the term refers to a linear combination of variables. The resulting function can be linear or curvilinear. All three of the polynomial functions below represent linear regression (note: we are ignoring a 0 degree case which is not useful in this context).
  1. y = ax + b (first degree, linear)
  2. y = ax2 + bx + c (second degree, quadratic)
  3. y = ax3 + bx2 + cx + d (third degree, cubic)
In the current project, we calculate sensor values using four reference standards (i.e., n = 4). Thus, a third-degree polynomial is the highest-degree polynomial that we can calculate. It is always the case that the highest-degree polynomial possible is n - 1, and in this case that means 3 (4 - 1).
Least-squares procedures are ordinarily used for linear regression. In this procedure, a line is fit such that the sum of the distances from each datum to the line is as small as possible. There are many programs available that use least squares procedures to perform linear regression. You can even use Excel (click here for more information).
It should also be noted that we do not have to use linear regression. We could use nonlinear regression. Examples of nonlinear regression result in a power function or a Fourier function. Linear regression, however, is well-suited to our project's data and, further, software correction (calibration) is easily implemented. In fact, in this project, I don’t believe you would gain much of anything by using nonlinear regression.
Choosing the Polynomial
In theory, we want to use the polynomial that best fits the data. That is, the polynomial that produces the smallest coefficient of determination, denoted r2 (or R2, pronounced “R squared”). The closer r2 is to 1, the better the fit. With least squares estimation, it is always the case that the higher the degree of polynomial used, the better the fit.
You do not, however, have to automatically use the highest-degree polynomial possible. Since calibration will take place in software, there may be cases in which the use of a lower-degree polynomial represents a speed and/or memory advantage, especially if the accuracy to be gained by using a higher-degree polynomial is very small.
Below, we demonstrate the calibration procedures for the HIH6030 sensor using polynomials of different degree, and in so doing we will illustrate the general procedure which is applicable to any degree of polynomial that you choose to use.
Using the data from the previous table, we first perform the least squares regression procedure to determine the coefficients for each polynomial. Those values will come from the regression software package used. The results are below, including the r2 values.
  1. Linear: y = ax + b; a = 1.0022287, b = -8.9105659, r2 = 0.9996498
  2. Quadratic: y = ax2 + bx + c; a = -0.0012638, b = 1.1484601, c = -12.0009745, r2 = 0.9999944
  3. Cubic: y = ax3 + bx2 + cx + d; a = 0.0000076, b = -2.4906103, c = 1.2061971, d = -12.7681425, r2 = 0.9999999
The observed values can now be modified using the calculated functions. That is, the sensor readings can be calibrated as illustrated in the table below (note that OBS, Corrected, and ERR values are rounded to two decimal places).
HIH6030 Observed and Calibrated Values Using Polynomials
  RAW   1st Degree   2nd Degree   3rd Degree
Ref RH OBS ERR Corrected ERR Corrected ERR Corrected ERR
11.30 20.79 9.49 11.93 0.63 11.36 0.06 11.30 0.00
32.78 40.77 7.99 31.95 -0.83 32.83 0.05 32.78 0.00
75.29 83.83 8.54 75.11 -0.18 75.85 0.55 75.29 0.00
84.34 93.43 9.09 84.73 0.39 84.83 0.49 84.34 0.00
RMSE   8.795736   0.562146   0.371478   0.00212
It can be seen that all three of the polynomials produced a significant decrease in the RMSE, compared to the observed measures, and that is why you calibrate. The graph below illustrates the improvement using the 1st degree polynomial. Note how the calibrated (corrected) data points now lie near the ideal diagonal.
Scatter plot of output from an HIH6030 sensor.
Calibration in Software
Once we have run the calibrations and chosen the polynomial, we can modify the software to incorporate the corrections into the sensor data. Using the HIH6030.c example, we can modify the program code as follows:
The initially calculated variable is RH. In the code lines above, we have created three new variables representing calibration (RHCal1, RHCal2, RHCal3). For illustration, we have created new variables using each of the three polynomials, whereas, in practice, you would likely calibrate the sensor value using only the chosen polynomial.
Summary Example
To summarize the steps for checking and calibrating a humidity sensor, a final example is presented using data from the DHT-22 (#2) sensor that was evaluated.
  • The first step is to evaluate the sensor performance using reference standards.
This was done using the salts in micro-environments to produce atmospheres with known RH. The data were collected and appear in a table previously presented. We can characterize sensor accuracy using the RMSE term, which also appears in the table. Based on the results of the first step, we can decide whether to perform sensor calibration. If so, then proceed to the second and third steps.
  • The second step is to perform linear regression to determine a function that relates the ideal RH value (from the standards) to the observed value from the sensor readings.
Here, we chose a cubic polynomial (y = ax3 + bx2 + cx + d, where y is the calibrated value and x is the sensor reading) and determined that the coefficients are a = 0.000091367, b = -0.01452993, c = 1.77623089, and d = -14.17403758.
  • The final step is to modify the sensor values using the polynomial to calculate the calibrated values.
The modifications (calibrations) are implemented in software and translate the sensor readings to their calibrated equivalents.
The graph below illustrates the result of the steps. The plot includes the observed values as well as the calibrated values (for the data collected) that are derived from the calibration polynomial.
Scatter plot of output from a DHT-22 sensor.
Closing Thoughts
In this project we demonstrated a method for evaluating the accuracy of common humidity sensors. That is, by utilizing chemical properties to develop reference standards, we can compare the sensor readings to standard values and independently determine the accuracy of the sensors. Furthermore, we demonstrated a software procedure for calibrating the sensor readings and thereby producing more accurate measures of relative humidity.
Sensor evaluation and calibration are procedures relevant anywhere that sensors are used. In the case of humidity sensors, this project demonstrates relatively inexpensive and easy methods of adding to their value.
8 Comments
  • D
    dimtass November 30, 2017
    Excellent article. Kudos!
  • AEKron November 30, 2017
    I never knew about the salts trick. I made a calibration chamber using the wet bulb/ dry bulb technique. This is an excellent article.
  • R
    Raymond Genovese December 02, 2017
    Thanks so much for the kind words.
  • joshctaylor December 08, 2017
    Good write up, the salts is a good wayto do it. A couple of notes about the regression - you should always use the lowest degree of polynomial to avoid over fitting the data. (linear means y=mx+c with no squares or higher orders). When calibrating sensors R^2 is of little use. The standard error , s,  is the important one. It is the standard deviation of the residual. This gives the accuracy of your calibrated sensor. 2 standard errors gives a confidence interval of 95 %. These sensors can also be sensitive to temperature so repeating the experiment in the fridge and somewhere hot will also give the temperature dependency.
    • R
      Raymond Genovese December 08, 2017
      First of all, thank you for the kind words and thanks for taking the time to respond. I’m not at all sure that I agree with you on restricting the regression to a first-order polynomial. I am not certain sure that I disagree in the sense that I would like to know the specifics of the rationale for that position. To me, in general, over-fitting is a consideration when either a) you have a large n (a lot of data) or b) you know a good deal about the underlying process. In the case of having many data points, it might be recommended that you back off of a 25th degree polynomial, for example. Some statisticians have even suggested to me that you look at the distribution of the r2 values and search for a large drop (as subjective as that might seem) and stop there. In the case of the underlying process argument: if your study involves some biological process, for example, you would be challenged to explain how that process could possibly produce a function resulting from a 25th degree polynomial (something along the lines of a simpler process is more likely). In the case of a humidity sensor and using 4 reference standards, I don’t see either as being an issue and pending more information (maybe something about the material behavior), I am fine with how I approached the issue. You are absolutely correct that the standard error of the regression can be used to assess the precision of the predictions and some definitely prefer that approach, period (see http://blog.minitab.com/blog/adventures-in-statistics-2/regression-analysis-how-to-interpret-s-the-standard-error-of-the-regression for example). I am fine using r2 to only measure the goodness of fit because, again, given the actual data, the r2 values are so large. If they were not, I would likely simply reject the sensor. For me, with regard to the humidity sensors, intuitively, r2=.99991 vs. r2=.99992 is acceptable and not legitimately discernible in many cases. That is not, however, measuring the precision of the prediction. I went back and looked at the 95% and 99% confidence intervals and was amazed at how small they get (e.g., in the last plot) when you increase the degree of the polynomial. But, you are right that I could have been more explicit on this point. When is “better” the enemy of “good”? Depends on what you are after I suppose. In preparation for this article I was struck with this paper (http://www.faslab.it/pdf/practical-rh-experiences.pdf) that demonstrates just how complex the issue can be in some real-world situations – you may enjoy the read. As for the temperature dependence, I don’t doubt that, but where is that dependence described to the point that you can make use of it? Your suggestion is sound, i.e., test it empirically. Thanks again.
  • Davor Predavec August 03, 2018
    Hi there,
    thanks for posting such a detailed article about sensors, I needed exactly that! But I encountered an error while compiling code for H8121.
    The error is:
    C:\Users\Davor\Desktop\8121\8121.ino:13:23: fatal error: qm_common.h: No such file or directory
    #include “qm_common.h”
    Could you help me with this, please?
    Many thanks!
    Davor
    • R
      Raymond Genovese August 03, 2018
      Thanks for the kind words. qm_common.h is a header file that is included in the D2000 software installation and you should have it if you have installed the board software for the D2000. It may be that you are trying to compile the code on an Arduino because I notice the .ino suffix in error message [8121.ino}. That code will not compile on an Arduino because it is for a D2000 and the I2C and terminal display code is specific for that board. The code simply reads the sensor and prints the reading so you should be able to find Arduino software/library to do that. Look around for code for the Honeywell sensors and if you get stuck after trying, you may want to ask the question in the Forum https://forum.allaboutcircuits.com/ Hope this helps.
  • S
    seanspotatobusiness May 27, 2020
    How long did your different sensors maintain their calibration? Don’t they tend to drift after a while? It would be good to know which of the plethora of sensors are stable enough to be worth going through the trouble. Only having to calibrate once every five years or similar would be nice. Thanks for the article, anyway.